Красноярск
Error No. 555 of AP ID 32
Meaning:
Basic unit: Unit was at one of the safety limit switches during operation. Emergency stop not pressed. Movements blocked.
Reason, Cause:
- Safety switch strips S80, S90 at foot end or S81, S91 at head end on support strut for table frame (collision protection between spotfilm device - table strut) actuated? - I.I. contact safety device safety switch S301 actuated? - Spotfilm device safety switch strip (collision protection on patient arm with tomo rod) actuated? - External safety circuit (ceiling photo-sensitive cell, door contact, step plate) open, X54 BUC jumper open? - A unit movement is at the safety limit switch (S150 tilt drive, S140 table lift, S160 and S161 table longitudinal). - Temperature on transformer too high > 105°? Check using SSW/ Monitoring/ Monitoring Status: M1 T1 thermal switch.
- SW-controlled BUC safety relays, K6, K7, deenergized? Reason: A movement took place without control or the potentiometer is defective. Watch for subsequent errors! - Chain break monitor for table lift responded, a microswitch, S6 - S9, actuated. - Temperature for table longitudinal drive too high? Switch S169 responded.
Action:
- Correct cause for possible collision. - Check the mechanical position of the movement to the electrically sensed position (potentiometer). SSW Diagnostic/ Tests/ Monitoring/ Monitoring: Unit position. - Using the SSW/ Diagnostic/ Tests/ Movement tests with Service buttons, move the drive out of the safety circuit. New potentiometer adjustment, possibly replace potentiometer (see the potentiometer Help text).
Этот сайт использует файлы cookies для более комфортной работы пользователя. Продолжая просмотр страниц сайта, вы соглашаетесь с Политикой использования файлов cookies.