siemens r200
kloun71Дата: Понедельник, 16.Фев.2015, 09:27 | Сообщение # 1
У вас сообщений: 273
тула
Возникла ошибка 555/035 при запуске аппарата, стол блокирован, помогите кто чем может.
anton26Дата: Понедельник, 16.Фев.2015, 11:28 | Сообщение # 2
У вас сообщений: 375
Error No. 555 of AP ID 32
Meaning:
Basic unit: Unit was at one of the safety limit switches during operation. Emergency stop not pressed. Movements blocked.
Reason, Cause:
- Safety switch strips S80, S90 at foot end or S81, S91 at head end on support strut for table frame (collision protection between spotfilm device - table strut) actuated?
- I.I. contact safety device safety switch S301 actuated?
- Spotfilm device safety switch strip (collision protection on patient arm with tomo rod) actuated?
- External safety circuit (ceiling photo-sensitive cell, door contact, step plate) open, X54 BUC jumper open?
- A unit movement is at the safety limit switch (S150 tilt drive, S140 table lift, S160 and S161 table longitudinal).
- Temperature on transformer too high > 105°? Check using SSW/ Monitoring/ Monitoring Status: M1 T1 thermal switch.
- SW-controlled BUC safety relays, K6, K7, deenergized? Reason: A movement took place without control or the potentiometer is defective. Watch for subsequent errors!
- Chain break monitor for table lift responded, a microswitch, S6 - S9, actuated.
- Temperature for table longitudinal drive too high? Switch S169 responded.
Action:
- Correct cause for possible collision.
- Check the mechanical position of the movement to the electrically sensed position (potentiometer). SSW Diagnostic/ Tests/ Monitoring/ Monitoring: Unit position.
- Using the SSW/ Diagnostic/ Tests/ Movement tests with Service buttons, move the drive out of the safety circuit. New potentiometer adjustment, possibly replace potentiometer (see the potentiometer Help text).
kloun71Дата: Понедельник, 16.Фев.2015, 15:11 | Сообщение # 3
тула
спасибо огромное
1. Feederadi[26.Дек.2024]
2. NovikovAN4[25.Дек.2024]
3. qutab[25.Дек.2024]
4. alayura[25.Дек.2024]
5. BR[25.Дек.2024]
6. zaharcenkodina[25.Дек.2024]
7. marker1k[24.Дек.2024]
8. Алексей4373[24.Дек.2024]
9. wowanww[24.Дек.2024]
10. traviolly[24.Дек.2024]