Здравствуйте уважаемые коллеги! Разрешите воспользоваться возможностью и попросить помочь в решении проблемы. Аппарат SIEMENS Axiom Iconos R200, ошибка 007/016 ,
Дата: Понедельник, 29.Янв.2018, 08:54 | Сообщение # 4
У вас сообщений: 165
ENGINEER
OFFлайн
Российская Федерация
USSR
Error No. 375 of AP ID 33
Meaning: Collimator: Limit switch error of the width collimator plates.
Reason, Cause: During initialization, the width collimator plates must interrupt the initialization photo-sensitive cell light beam within the predefined time (search run position). If the photo-sensitive cell light beam does not change its status within this time period, a timeout occurs. Probably a mechanical problem has occurred.
Action: - Are width collimator plates mechanically jamming (transport safety device!)? - Is the U3 photo-sensitive cell on the D1 board defective? Check the V7 LED for the U3 photo-sensitive cell on the D1 board. Replace the D1 board. - SM2 width drive failed? - Replace the collimator.
Советую софт на компе переустановить, сохранив на всякий случай текущий образ. Главное файл лицензии и бэкапы настоек найдите и себе сохраните прежде чем что то делать! А фото для чего обрезали?:(