Somatom AR error EX2842
janekДата: Понедельник, 21.Ноя.2011, 16:41 | Сообщение # 1
Автор темы
У вас сообщений: 108
инженер-электроник
москва
Здравствуйте коллеги.На удаленном обслуживании томограф Сименс Somatom AR.Документации под рукой нет,надо ехать.Позвонили ,при включении высокого томограф уходит в ресет,перезагрузки не помогли,надо ехать.Что это может быть.ошибка EX2842,трубка ,генератор,стартер?
ЭлектрослесарьДата: Понедельник, 21.Ноя.2011, 21:54 | Сообщение # 2
У вас сообщений: 43
электрослесарь
деревня Мурино
CODE: EX2842 SHORTMESSG: TDR_TIME_ERR
DETAILED DESCRIPTION:
Time from XCENA to TDR_OK was longer than specified in table T55. The signal _
XCENA is set from the DMS function controller ROSY to start high tension and
radiation; the signal _TDR_OK is set by the XRS function controller when the tube current and the high tension are in range. Please see also the trouble
shooting guide in the section of 5 of the ACS manual.
Reason for error:
- Bad values in table T55 or T56.
- Faulty inverter D22.
- Faulty tank board D28.
- Faulty control board D27.
- Wrong polarity in cable between inverter and ROTANX (only after change).
- Bad adaption values for filament.
- Faulty Rotanx.
What to do?
1:
Check the timeout value in table T55 (Row 00, byte 3) with SERVICE/
ANALYZE. It should not be less than 15 ms (0F hex).
2:
Check the kV, mA and filament values in table T56 with SERVIC/ANALYZE. If
any are not correct, perform a new "INIT" to load the tables to the ACS. It
may be possible to load the tune up tables from tape and also perform a new
- kV should be 00-A8 hex (50-134 kV).
- mA should be 26-7D hex (38-125mA).
- Filament should be 1F0-2A0 hex.
3:
If the error occurs after a change of single tank ROTANX, check the
inverter connection W216 from inverter D22 to Rotanx, X224 AC1 - X252.2.
and X224 AC2 - X252.1. This cable X229-X252 should be twisted and laid as
close as possible to the Rotanx.
4:
Check the other connection as well:
- W217 from tank board D28 to control board D27, X243-X242. The shielding of
this cable should be connected to ground at both ends. Furthermore it should
be laid as far away as possible from the cable W216.
- Cable from the Rotanx to the tank board D28, -X220.
5:
Measure MANEG and MAPOS at the tank board D28 during a scan in static mode.
6:
6 upto 8 if the result of the measurement is not correct:
- In the case of replacing D27; D27 replaces, in old systems, both the CPU
board D26 and the I/O board D27.
The value of MANEG increases continuously during the scan, change the
polarity of the cable W216 (X224 AC1 - X252.2, X224 AC2 - X252.1) at the
connection X252/Rotanx. This is only possible after inverter - or ROTANX
exchange. Perform a cold start or an "INIT" to load the generator tables
again.
7:
The value of MANEG is too low:
- Check th cable W315 from the control board D27 to the inverter D22, X236-
X237.
- Check the switch signals CH1 and CH2, if missing replace control board D27.
- Check the inverter current with a current transformer of current probe.
- Is the system working with maximum pulse width (closed to 50usec) replace the
single tank ROTANX.
- Is one side of the inverter current missing (positive or negative wave)
replace the inverter D22.
- Replace tank board D28.
8:
The value of MANEG is too high:
- Replace control board D27.
9:
Measure HTPOS and HTNEG at the tank board D28 during a scan in static mode.
10:
10 upto 12 if the result of the measurement is not correct:
-.In the case of replacing D27; D27 replaces, in old systems, both the CPU
board D26 and the I/O board D27).
The value of HTPOS and HTNEG are too low:
- Check the cable W315 from the control board, D27 to the inverter D22, X236-
X237.
- Check the inverter current with a current transformer or current probe.
- Is the system working with maximum pulse width (closed to 50 usec) replace
the single tank ROTANX.
11:
The values of HTPOS and HTNEG are too high:
12:
The value of MANEG and MAPOS and the value of HTNEG and HTPOS is range (
reaches selected value within specified time from table T55):
- Check grounding of tank board D28, -X265. This cable should be laid as far
away as possible from the cable W216 (inverter D22 - ROTANX).
- Check cable W216 (inverter D22 - Rotanx) and W217 (tank board D28 - control
board D27).
ЭлектрослесарьДата: Понедельник, 21.Ноя.2011, 21:57 | Сообщение # 3
У вас сообщений: 43
электрослесарь
деревня Мурино
Сбрось адрес - пришлю всё в электронном виде!!!
janekДата: Понедельник, 21.Ноя.2011, 23:22 | Сообщение # 4
Автор темы
инженер-электроник
москва
Спасибо,список неисправностей внушает дрожж.Я с Сименсами мало общался,но надо ковыряться.Мой адрес kosh47@mail.ru .Как понял кирдык трубе не исключен.
1. УЗИ Philips HD3 EXP[VG54 (13.Авг.2025)]
2. Электрон КРТ стол SUP...[Benq (13.Авг.2025)]
3. Размораживатель плазм...[yblod5 (13.Авг.2025)]
4. Диаметры трубок для а...[ALROS (12.Авг.2025)]
5. Термостат ТС-80М-2[yblod5 (09.Авг.2025)]
6. Радиовизиограф Woson[seabee (09.Авг.2025)]
7. Скалер Piezotome Solo...[seabee (09.Авг.2025)]
8. Стерилизатор паровой ...[fiksa555 (07.Авг.2025)]
9. Проблемы с ПО "И...[Long (07.Авг.2025)]
1. Philips HD3-EXP[13.Авг.2025]
2. Рентгеновское оборудо...[08.Авг.2025]
3. VersaMed iVent201 Сер...[13.Фев.2025]
4. Гелпик КРД Ренекс РЦ ...[09.Фев.2025]
5. Офтальмологическая ус...[31.Окт.2024]
6. Cardinal-Health-VELA-...[03.Окт.2024]
7. Service Manual HAMILT...[02.Окт.2024]
8. Mindray DC-7 (ПО)[09.Сен.2024]
9. Руководство по эксплу...[26.Авг.2024]
10. Пароли для принтеров ...[20.Авг.2024]
1. dimcik[13.Авг.2025]
2. hiksus[13.Авг.2025]
3. seregond[13.Авг.2025]
4. Soldier[12.Авг.2025]
5. Foxsi[12.Авг.2025]
6. Farius[12.Авг.2025]
7. mirZAY[12.Авг.2025]
8. oemoonco[12.Авг.2025]
9. osubratsk[12.Авг.2025]
10. Alfscreen[12.Авг.2025]
Этот сайт использует файлы cookies для более комфортной работы пользователя. Продолжая просмотр страниц сайта, вы соглашаетесь с Политикой использования файлов cookies.
